ORIGINAL CONTRIBUTION
    
PDF of Article    Email Article 

Home
Journal
Now in Medline / Index Medicus & ISI



Subscribe
Search
Discussion Forums
Join
Multimedia
Meeting Highlights
Lecture Presentations
Surgical Video Library
Private Lecture Series
Audio Interviews
Resources
Medline
Learning Center
Meetings
For Authors
Editorial Board
Sponsors

Outcomes: The Key West Meeting


World Society of Cardio-Thoracic Surgeons


5th Congress Update in Cardiology and Cardiovascular Surgery
Sept. 24 - 28, 2009


Boehm.JPG: Contrast-Enhanced Magnetic Resonance Angiography for Control of Minimally Invasive Coronary Artery Bypass Conduits (MIDCAB/OPCAB)

(#1999-4336 ... June 25, 1998)

Dieter H. Boehm, MD, PhD1, Bernd J. Wintersperger, MD2, Hermann Reichenspurner, MD, PhD1, Helmut Gulbins, MD1, C. Detter, MD1, F. Kur, MD1, Bruno Meiser, MD1, Bruno Reichart, MD1

University of Munich, Klinikum Grosshadern, Munich, Germany
1Department of Cardiac Surgery
2Diagnostic Radiology

ABSTRACT

Objective: The purpose of this study was to delineate the course and determine the patency of venous and arterial conduits in the early postoperative period following minimally invasive bypass grafting. A less invasive magnetic resonance angiogram was evaluated as alternative to standard contrast angiography and cardiac catheterization.

Methods: Twelve patients (8 males and 4 females) with a mean age of 65.3 (± 7.4 ) years were evaluated four to seven days following minimally invasive direct coronary artery bypass surgery (MIDCAB) or off-pump multivessel revascularization with the Octopus™ stabilizer on the beating heart. Altogether 17 coronary bypass grafts were investigated: 12 left-sided mammary artery grafts to the LAD and five aortocoronary venous bypass grafts. The examination was performed with a 1.5 Tesla Magnetom Vision (Siemens AG, Erlangen) with phased array coil technology. Data acquisition was done with an ultrafast 3D gradient-echosequence in single breathhold and sagittal and coronal views. Contrast enhancement of the vessels was performed with automatic intravenous bolus injection of Gadolinium-DTPA after determination of the individual contrast transit time. Traditional contrast angiography was obtained in all patients during the same time period as a comparison to assess the sensitivity and specificity of the magnetic resononance imaging.

Results: All five venous grafts and 11 of the 12 IMA grafts were detected and shown to be patent with the MRA technique. Contrast angiography demonstrated complete patency for all 17 bypass grafts with adequate anastomoses and no evidence of stenosis. The calculated sensitivity for the visualization with MRA was therefore 92% for IMA grafts and 100% for venous grafts.

Conclusion: The contrast-enhanced ultrafast MRA in single breathhold technique is a reliable, noninvasive method for visualization and determination of the patency of arterial and venous coronary grafts.

INTRODUCTION

Diagnostic investigation of previously constructed coronary artery bypass grafts is routinely performed by means of injected radio-opaque contrast with cineangiography. Standard contrast angiograms also can be coupled at the same sitting with graft interventions like angioplasty or stenting if needed. However, coronary angiography uses ionic radiation, is expensive and has potential hazards. Contrast enhanced magnetic resonance angiography (MRA) is a recently implemented, noninvasive and fast imaging modality, although it needs some technical requirements. Especially for less invasive methods of coronary revascularization, the method would allow a less invasive assessment of graft status and assist in quality control. This report compares early postoperative MRA against standard coronary angiography to determine if the sensitivity and specificity of MRA can reach a level suitable for postoperative assessment of graft status.

MATERIALS AND METHODS

The examination was performed with a 1.5 Tesla Magnetom Vision™ (Siemens Medical Systems, Erlangen, Germany) with phased array coil technology. Data acquisition was done with an ultrafast 3D gradient-echosequence in single breathhold with sagittal and coronal views. Contrast enhancement of the vessels was performed with automatic intravenous bolus injection of Gadolinium-DTPA (Magnevist, Schering AG, Berlin, Germany) after determination of the individual contrast transit time. Data aquisition time was 23 and 29 seconds (sagittal and coronal views, respectively), and spatial resolution was 0.98x1.40x1.60mm3. The complete examination required a time exposure of 20 to 25 minutes. A consecutive 3D maximum intensity projection (MIP) calculation of the datasets was also performed.

Image evaluation was performed assessing contrast enhancement, pulsation, and breathing artifacts, as well as wrap-around artifacts using a five-step scale. For patency assessment, grafts were divided into three parts: proximal anastomosis, graft course, and distal anastomosis. The mean score of image quality was good (2.2 ± 0.8) and all MR datasets except one were of sufficient image quality for assessment of bypass graft patency. Subsequently the grafts were assigned to different groups: visible (which means patent) and non-visible (which means occluded). For purposes of correlation, the MR images were compared to a standard contrast angiography performed in the same patients during the same time period.

Twelve patients (8 males and 4 females) with a mean age of 65.3 years were evaluated four to seven days following off-pump coronary grafting. A total of 17 coronary bypass grafts were investigated with both techniques [see Table 1:592:]: 12 left internal mammary artery grafts to the LAD and five aortocoronary venous bypass grafts were performed. Seven patients received a single bypass using a LIMA graft to the LAD. The surgical approach in these cases was via a left-sided anterolateral mini-thoracotomy and cardiac stabilization with the CTS pressure stabilizing system (Cardiothoracic Systems, Cupertino, USA). In five patients, in addition to the IMA graft to the LAD, an aortocoronary venous graft was performed: in three cases to the RCA, and in two cases to the first diagonal branch. In these patients, the surgical approach was via median sternotomy and cardiac stabilization was achieved using the Octopus™ suction system (Medtronic GmbH, Düsseldorf, Germany).

RESULTS

All five saphenous vein grafts (SVGs) and 11 of the 12 LIMA grafts were shown to be patent using the MRA technique, resulting in a patency rate of 92% for IMA grafts and 100% for SVGs. Coronary angiography demonstrated full patency for all 17 bypass grafts [see Table 2 :596:]. Sensitivity for the visualization of IMA grafts with MRA was thus 92% and 100 % for SVGs. Examples of a MRA visualization of a LIMA to LAD graft and the corresponding angiography are shown in Figure 1a and Figure 1b. A saphenous vein graft in addition to a LIMA to LAD graft in MRA technique is shown in Figure 2. Failure of MRA in detecting the one IMA graft shown to be patent by coronary angiography was due to inadequate patient compliance with the breathhold technique.

DISCUSSION

Visualization of venous bypass grafts by means of MR imaging was reported by Herfkens in 1983 [Herfkens 1983]. Subsequent studies for assessment of IMA grafts were performed in 1987 but without comparing the results to established methods [Gomes 1987]. MR flow studies of coronary artery bypass grafts with assessment of the cyclic flow profile were published by Hoogendoorn et al. in 1995 using phase contrast MR techniques [Hoogendoorn 1995].

The contrast-enhanced ultrafast MRA in single breathhold technique is a fast, reliable and noninvasive method for visualization and determination of the patency of coronary grafts after MIDCAB. Due to the high spatial resolution, both venous and arterial grafts can be visualized. The main limitations of this method are: 1) adequate patient compliance (breathholding technique), and 2) exclusion of patients with permanently implanted pacemakers or defibrillators. Functional assessment and quantification of bypass-stenoses is currently not possible with this technique. In the future the combination of MRA with noninvasive MR flow measurements, making use of phase contrast technique, could demonstrate flow profiles. To allow quantification of IMA flow rates, however, further improvements in spatial resolution are necessary.

The advantage of MRA is a short procedure time, generally in the range of 20 to 25 minutes. Expensive ECG-triggering is not necessary for simple bypass visualization or for a 3D dataset. There are numerous possibilities for postprocessing and visualization of the entire bypass course [Manning 1993].

The disadvantage of MRA is the expense of the paramagnetic contrast medium. However, Gadolinium-DTPA has less side effects than iodine based injectable contrast medium. The evaluation of the distal part of the anastomosis, however, is not possible with the current magnetic resonance technique. MRA examination is very dependent on the collaboration between patient and radiologist to avoid artifacts caused by inconsistencies with the breathholding technique. Today, MR-angiography is available only in a few medical centers having MRT equipment and additional expensive high-performance gradient coil systems.

Contrast enhanced MRA obtained a sensitivity of 92% for the visualization of IMA grafts and 100% for visualization of saphenous vein grafts in this study, which is superior to other MRT techniques. Thus contrast enhanced 3D magnetic resonance angiography has the potential for being a reliable method for CABG visualization and patency assessment in the early postoperative period. Three dimensional postprocessing is helpful in delineating the course of grafts. For further evaluation of occluded grafts or for angioplasty, conventional angiograms are still necessary. X-ray angiography is invasive and can be associated with major complications. In comparison, contrast enhanced MR angiography is a noninvasive modality without iodine contrast medium or ionizing irradiation, and can be performed in less than 30 minutes with minimal risks.

AUTHOR/ARTICLE INFORMATION

Presented at the First Annual Meeting of the International Society for Minimally Invasive Cardiac Surgery, Minneapolis, Minnesota, June 25-27, 1998.

Reprint requests to: Dieter H. Boehm, MD, PhD, Department of Cardiac Surgery, University Hospital Munich-Grosshadern, Marchioninistr. 15, D 81366 Munich, Germany; Phone: +49/89/7095-3464; Fax: +49/89/7095-8873; E-mail: boehm@hch.med.uni-muenchen.de

Submitted on: Peer reviewed and accepted at the International Society for Minimally Invasive Cardiac Surgery's 1st Annual Meeting and Scientific Sessions, Minneapolis, Minnesota, June 25-27, 1998.

Keywords: MRI, MRA, MIDCAB, OPCAB, coronary bypass grafts; Gd-DTPA

REFERENCES

1. Gomes AS, Lois JF, Drinkwater DC Jr, Corday SR. Coronary artery bypass grafts: visualization with MR imaging. Radiology 162:175-9, 1987. :3491379:

2. Herfkens RJ, Higgins CB, Hricak H, Lipton MJ , Crooks LE , Lanzer P, et al. Nuclear magnetic resonance imaging of the cardiovascular system: normal and pathologic findings. Radiology 147:749-59, 1983. :6601813:

3. Hoogendoorn LI, Pattynama PM, Buis B, van der Geest RJ , van der Wall EE , de Roos A. Noninvasive evaluation of aortocoronary bypass grafts with magnetic resonance flow mapping. Am J Cardiol 75:845-8, 1995. :7717297:

4. Manning WJ, Li W, Edelmann RR. A preliminary report comparing magnetic resonance coronary angiography with conventional angiography. N Engl J Med 328:828-32, 1993. :8285929:

 


ISSN#: 1522-6662
Copyright 2009 Forum Multimedia Publishing, LLC. All rights reserved.

The material available at this site is for educational purposes only and is NOT intended for any diagnostic, clinically related, or other purpose. Forum Multimedia Publishing, LLC, assumes no responsibility for any use or misuse of this material and makes no warranty or representation of any kind with respect to the material available at this site.