Clinical Experience with the CorLink Device for Proximal Anastomosis of the Saphenous Vein to the Aorta: A Clinical, Prospective, and Randomized Study
(#2002-71002 ... September 25, 2002)
Friedrich-Christian Riess, MD,1 Hanns Helmold, MD,1 Irene Hilfer, MD,1
Ralf Bader, MD,1 Jan Stripling, MD,1 Christine Loewer, MD,2 Frank Wesner, MD,3
Niels Bleese, MD1
Departments of 1Cardiac Surgery and 2Cardiac Anesthesia, Albertinen-Krankenhaus;
3Praxis Fleethof, Hamburg, Germany
View supplemental Quicktime movie
for this article: Movie 1
Movie 2
Movie 3
ABSTRACT
Background: Avoiding tangential clamping of the
ascending aorta during coronary bypass operations reduces
the trauma to the aorta and may avoid local particulate
embolization.
Methods: From December 2000 to May 2001, 21 male
patients, mean age 64.1 ± 7.2 years (range, 46-76 years), with
coronary artery 2-vessel (n = 3) and 3-vessel (n = 18) disease
were divided randomly into 2 groups and underwent myocardial
revascularization. In 11 patients an aorta-saphenous vein
graft anastomosis was performed with the CorLink device for
anastomosis between the saphenous vein and the ascending
aorta. Ten patients served as control subjects. In these
patients the central bypass anastomosis was performed with a
6-0 running suture. Clinical follow-up was performed
1 month and 3 months postoperatively. Six months after
surgery, multislice computed tomography was performed to
evaluate bypass patency for all patients.
Results: Mean number of study vessels was 1.2 ± 0.4 in the
CorLink group and 1.5 ± 0.5 in the control group. In the CorLink group, 13 additional arterial and vein grafts were performed,
and in the suture control group 15 additional mammary
artery grafts were carried out. No intraoperative complications
occurred. In 2 CorLink anastomoses an additional
stitch was necessary because of minor bleeding. Follow-up was
carried out at 6 months with multislice computed tomography
for all patients and showed only 1 study vessel occlusion in the
CorLink group. All 62 other bypass grafts were revealed to be
patent and had anastomoses of good quality.
Conclusion: Our experience suggests that the CorLink
device is a safe and effective technique for anastomosis
between saphenous vein grafts and the ascending aorta. The
CorLink device could be used for totally endoscopic coronary
bypass operations. Further randomized studies enrolling a
larger number of patients are necessary to determine which
patients may benefit the most from this procedure.
 Click here for a PDF version of the full article. (Subscribers Only)
|
 |