Six-Month Angiographic Evaluation of Beating-Heart Coronary Arterial Graft Interrupted Anastomoses Using the Coalescent U-CLIP Anastomotic Device: A Prospective Clinical Study
(#2002-77117 ... September 23, 2002)
Michael P. Caskey, MD,1 Merick S. Kirshner, MD,1 Edwin L. Alderman, MD,2
Sonna Lea Hunsley, RN, BS,1 Michael A. Daniel, MS, MBA3
1St. Joseph's Hospital, Phoenix, Arizona; 2Stanford University Medical Center, Palo Alto,
California; 3Daniel and Daniel Consulting, Orinda, California, USA
View supplemental Quicktime movie
for this article: Movie 1
Movie 2
Movie 3
ABSTRACT
Background: Interrupted suture technique avoids the
"purse string" and puckering effects frequently seen with continuous
suture techniques and should represent the standard
of care in the creation of high-quality vascular anastomoses.
This clinical study evaluated the safety and effectiveness of a
self-closing surgical clip (Coalescent Surgical U-CLIP Anastomotic
Device [U-CLIP]) designed to facilitate this interrupted
technique. Left internal mammary artery (LIMA) to
left anterior descending (LAD) coronary bypass grafting
was studied.
Methods: Eighteen patients meeting inclusion criteria
were enrolled (October 2000 through September 2001) into
this prospective study. Anastomoses were performed using a
beating-heart median sternotomy procedure (off-pump
coronary artery bypass) in 17 cases (94%) and a minimally
invasive beating-heart procedure (minimally invasive direct
coronary artery bypass [MIDCAB]) in one case (6%). Sixmonth
follow-up was completed on 18 patients (100%), with
angiograms performed on 17 patients (94%) at a mean of
179 days (range, 168-191 days). Qualitative and quantitative
angiographic assessment was performed by an independent
core laboratory.
Results: The U-CLIP was used for 18 LIMA-to-LAD
interrupted anastomoses without the requirement for knot
tying or suture management and with no device-related morbidity
or mortality. Mean LIMA-to-LAD anastomosis time
was 8.6 minutes (range, 5-14 minutes). All anastomoses were
FitzGibbon grade A at 6 months postprocedure. Quantitative
analysis showed mean luminal diameters proximal to the
anastomosis of 2.32 mm, at the anastomosis of 2.25 mm, and
immediately distal to anastomosis of 1.99 mm. The average
ratio of anastomosis to LAD diameter was 1.17 (range, 0.93-
1.93). Anastomotic stenosis as a percentage of average
LIMA/LAD diameter was a negative 4.2%, comparing favorably
with the 23% to 24% reported in the POEM (Patency,
Outcomes, Economics of MIDCAB) study.
Conclusions: The interrupted technique, facilitated by a
self-closing anastomotic clip, yielded 6-month follow-up and
angiographic results that compared favorably with results of
other published studies.